we are using Agilent 7000b MSD and we are using the method given in Agilent pestanalyzer.In that last oven program temperature and transferline temperature is 280 Source temperature is 300 and quad temperature is 180.Is this cause this any problem in sensitivity.
You can do a test by running the same sample with a source setting of 150C, 200C, 250C and 300C and compare the sensitivity to see what the effect of temperature is on your specific analysis. Hotter temps can keep the source cleaner, but I have also found that I see higher noise and background when running higher source temps.