Advertisement

Impurity peak in WORKING STANDARD

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
I see an impurity peak in the working standard. System suit and all other criteria pass. What is the limit of imurities that I can ignore or what is the maximum area of the impurity that can be disregarded according to USP?

Thanks in advance
What does your SOP say? :twisted:

Ph. Eur. suggests a general disregard limit of 0.05% (chapter 2.2.46) if I recall correctly.
Thanks for the reply. SOP does not comment anything on impurity in WOrking standard. However it does limit amount of impurity in the sample.
3 posts Page 1 of 1

Who is online

In total there are 247 users online :: 1 registered, 0 hidden and 246 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: Google [Bot] and 246 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry