If your internal standard give a peak size in the middle of the range of your analytes in the calibration curve, you should be good.
Tell more about the analysisi - if you can, what the analytes are, at least compound classes. Tell how you made the standards serial dilution? All diluted directly down from a single solution?
How many data points across the peak? Do you switch between one set of transitions to another during the run or do you use the full set the whole way through? And if you have timed events switching sets of transitions, how close is the switch to the starts and ends of peaks?
Thank you for your response.
Sorry for forgetting to tell you that I am doing with organophosphorus pesticide.
And all calibration solutions came from the same stock solution.
I make 10 ppb, 20 ppb, 50 ppb, 100ppb, 150 ppb and 200 ppb as conentration for calibration.
The data points are keep more than 12 per peaks.
And I cut the transitions into several time interval, but as I am using a 50 mm column, there are some overlapping between transitions.
I am sorry that I do not understand what is time events switch of transitions? Does it mean the ions scan rate?
I tried to cut the number of transition and isolated transition of internal standard from other transition.
Half of the result give better R^2 (at least around 0.99)
And the other sets seems contain signal saturation problem.
The curve of first 3 points are linear, and last 3 points are also linear (but lower slope), but with combining them into 6 points calibration, it gives a nonlinear curve.
I would tried to reduce either the capillary voltage or collision voltage of each transitions to see if it helps. (coz I do not want to only have working range from 10ppb to 50 ppb )