temperature settings 5973
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
5 posts
Page 1 of 1
I vented the system last week and cleaned the source, during pump down the temperature (actual) reading for the source heater was bouncing around between 170-500 degrees (set at 230) and the quadrupole never reached the set temperature of 150 (stayed around 80). I was told main board or side board by agilent. Any ideas or suggestions in case its something I can try to reset. I have already redone the pump down after opening the system back up and checking connections and changed out the heater source assembly.
-
- Posts: 3477
- Joined: Mon Jan 07, 2013 8:54 pm
If the heaters are new replacements, then remove the covers on the side board and find the pass through connectors and see if there is voltage being applied to the heater elements when it should be heating.I vented the system last week and cleaned the source, during pump down the temperature (actual) reading for the source heater was bouncing around between 170-500 degrees (set at 230) and the quadrupole never reached the set temperature of 150 (stayed around 80). I was told main board or side board by agilent. Any ideas or suggestions in case its something I can try to reset. I have already redone the pump down after opening the system back up and checking connections and changed out the heater source assembly.
Was the temperature fluctuation in the source rapid or over several hours? If rapid I would suspect a problem with the thermocouple or the circuitry that reads the temperatures from it.
The past is there to guide us into the future, not to dwell in.
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
It sounds very much as if you did not get a good connection when you reconnected the heater sensor. Opening up and doing it again will surely work out a lot cheaper than a new circuit board.
Peter
Peter
Peter Apps
-
- Posts: 158
- Joined: Wed Mar 05, 2008 2:56 pm
Hi
As suggested check the thermocouple and heaters first, they do break on 5973s and can be replaced. Also check the ribbon cable that runs from the side board to the mainboard.
As suggested check the thermocouple and heaters first, they do break on 5973s and can be replaced. Also check the ribbon cable that runs from the side board to the mainboard.
-
- Posts: 4
- Joined: Tue Aug 28, 2012 11:47 am
If the temperature is fluctuating rapidly, I would bet that it is the thermocouple is shot. I've seen it happen several times.
5 posts
Page 1 of 1
Who is online
In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Baidu [Spider] and 1 guest
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Baidu [Spider] and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science