I'm developing a new HPLC method and I noticed that as I was entering the injection amount I saw an option for "Standard Injection" or "Injection with Needle Wash." What is the difference between these two? Is there a preference?

B

Re: Standard wash vs Needle

Posted: Thu Sep 12, 2013 2:25 pm
by Consumer Products Guy
On the Agilent 1100, needle wash actually dips the needle into a designated autosampler vial containing wash solvent. This option is there for those who wish to rinse the outside of the injector needle (the inside is constantly being flushed with mobile phase) except during the injection cycle).

What autosampler do you have?

Re: Standard wash vs Needle

Posted: Thu Sep 12, 2013 3:48 pm
by Rob Burgess
The problem is that issues with carryover on the 1100 are most often associated with needle seat contamination issues which makes the needle wash virtually redundant. I prefer to run with it off.

Re: Standard wash vs Needle

Posted: Fri Sep 13, 2013 10:10 am
by LCbob
Hi

Also to add, if you do use the needle wash remember not to have a cap on the vial and to regularly change the wash solution. Overall my experience is the these two things don't happen.

Re: Standard wash vs Needle

Posted: Sat Sep 14, 2013 12:08 am
by Consumer Products Guy
We use Agilent 1100, and would rarely, if ever, use needle wash.

Re: Standard wash vs Needle

Posted: Mon Sep 30, 2013 1:16 pm
by brendandamm
Thank You!!!

Re: Standard wash vs Needle

Posted: Thu Oct 03, 2013 8:41 am
by lmh
hm, I personally use the needle wash. What does it cost? One vial, and 30 sec to set it up. I can't say I've ever looked in detail to see if it's necessary though. And LCbob, I never use a cap on the vial, and I replace the wash vial every sequence, without fail. You're right, there's no point otherwise.

Re: Standard wash vs Needle

Posted: Wed Oct 09, 2013 8:53 pm
by gerapas
could you explain why the vial must not be capped? Thanks