Hi
At presently I have just joined analytical method development,
I am confused how to calculate capacity factor practically.

Thank you in advance
Ashvin

1. Measure the retention time of your peak and the dead time of your column.
2. Divide retention time by dead time and subtract 1

k' = (tR/t0) - 1
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

Thanx ,
But how to calculate columne dead volume?
Ashvin
Ashvin Panchal
How to calculate column dead time
Ashvin Panchal

You don't calculate it, you measure it, just like you measure retention time. If you don't see a "solvent front" on your chromatogram, try injecting pure acetonitrile or methanol (I'm assuming you're using reversed phase here) and look for a small disturbance on the baseline.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
5 posts Page 1 of 1

Who is online

In total there are 179 users online :: 0 registered, 0 hidden and 179 guests (based on users active over the past 5 minutes)
Most users ever online was 10230 on Thu Dec 04, 2025 5:56 am

Users browsing this forum: No registered users and 179 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry