CD 25A resetting itself

Discussions about IC and related topics

2 posts Page 1 of 1
I have an old DX 600 system with a CD 25A that had been performing well until recently. it now turns off the power to the modulle whenever I try to set the suppressor amps either at the module or from the software. I have changed all of the cards in the module with known good cards, replaced all of the cables and changed out the suppressor with no luck. Has anyone ever seen this? Dionex told me that this was not seen by them before. Thanks for your help.
Hi

I suspect the problem is with your suppressor power supply, a short could be pulling down the Main Power Supply in the CD25A module. If this is the case you may have to speak with the local Thermofisher Ion Chromatography Service people as these units have limited parts supply.

Cheers

Mark Albertson
Sales Manager ICSP/HPLC
Thermofisher Australia
2 posts Page 1 of 1

Who is online

In total there are 105 users online :: 0 registered, 0 hidden and 105 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 105 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry