SPE cartridge for polyethyleneimine (PEI)

Posted: Tue Sep 10, 2013 6:32 pm
by JI2002
I'm trying to find a SPE method for PEI. The sample matrix is a serum-free tissue culture media. We know that there are several amino acids present in the sample (confirmed by LCMS), but there may be other unknown components. The goal is to remove the sample matrix by using a SPE cartridge. Any idea about the types of cartridges and the conditions for this application is greatly appreciated.

Re: SPE cartridge for polyethyleneimine (PEI)

Posted: Tue Sep 10, 2013 11:00 pm
by Don Shelly
I'm trying to find a SPE method for PEI. The sample matrix is a serum-free tissue culture media. We know that there are several amino acids present in the sample (confirmed by LCMS), but there may be other unknown components. The goal is to remove the sample matrix by using a SPE cartridge. Any idea about the types of cartridges and the conditions for this application is greatly appreciated.

Use a weak cation exchanger like a carboxylic acid sorbent.
Adjust sample pH to 7 with a volatile buffer or acid/base
pass through cartridge at 1 -3 ml per minute
wash away interferences with solvents
elute with methanol adjusted to pH 2.5 or lower at 1 ml per minute

Re: SPE cartridge for polyethyleneimine (PEI)

Posted: Wed Sep 11, 2013 3:29 pm
by JI2002
Thank you very much, Don. I'll give it a try.