Hi there, we are struggling a bit to explain differences in tailing factor we have seen for these two instruments for a simple fast dissolution LC assay. On an Agilent system (tested on 3 different binary systems) we saw tailing factors between 1.8-1.9, but when transferring the same samples, column and mobile phase onto an Alliance we saw TF's of between 1.4-1.5. Our method SS limits were specified as 0.8 to 1.8 so we are failing in some instances.

I seem to remember many years ago I came across a similar incident between these two instruments but couldnt explain it then either. This is the first time in a while I have seen this. Is there something with the injection mechanism differences between the two instuments that would cause such a seemingly drastic difference?
Hi Rob,

I don't know if I can explain the Tf value differences either, but here's a run-down of the Alliance 2690/5 injector in contrast to what I recall about the Agilent's: (would help if I knew if the binary systems you speak of were 1100 or 1200 systems.)

Side-port needle. (The Aglient has a straight-thru needle.)
Seal Wash and Needle Wash. (Agilent has only a Needle Wash, washing mechanisms differ between the two vendors.)
Flow-Through Needle Design, sample loop is integrated into the eluent flow path. (Can't recall at the moment, but seem to remember that the quaternary gradient and binary gradient LCs Agilent sells may have either Flow-Through Needles or a Fixed Loop design where the loop is separate from the eluent flow path.)

The only guess I have is to check into the nature of the needle wash solvents used on both systems...as well as to understand exactly what is done when the needle is washed by the autosampler for both types of LC. This may be the source of the problem...or differences in the rise times for each instrument...longer rise times can be manifest in larger values of Tf, though your case would push this guess to an extreme such that it isn't that likely to be the cause.

Otherwise, unless the Agilent and/or Waters LCs were modified, another thing would be that the capillary tubing ID on the Alliance systems would be larger than those for the Agilents, 0.009" versus 0.007" in the standard configurations for each system.
MattM
2 posts Page 1 of 1

Who is online

In total there are 5 users online :: 4 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Baidu [Spider], Bing [Bot], Semrush [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry