OK, I'll put my hands up: I have no idea whether it's possible to use batch export to export manually-adjusted baselines. If you really need to use manual integration, I may have answered the wrong question (sorry!).
But I'd suggest you look into what Chemstation can do: Firstly, it may be that you don't need manual integration. Have you tried the different sorts of baseline that Chemstation offers (advanced, classic, no penetration etc.)? Adjusting these, and choosing good integration parameters might avoid the need for manual alterations (good starting-conditions can be obtained by choosing a weak signal and using autointegrate). Note that you can set individual integration parameters for each signal of interest (or change entire defaults for a detector if you prefer). The exception is the baseline; if you choose "advanced", then everything will be "advanced" (usually no bad thing!).
There was a lengthy debate here a while back on the subject of manual integration, and its acceptability. As I remember, we agreed to differ, but the consensus was manual is OK, but if you can use automatic, it raises less awkward questions about validation, as well as being easier. The old hands reckoned they can do a better job than automatic; us newbies reckoned we like our automatic and if a chromatogram is so bad we can't get it to integrate automatically, it probably isn't worth manual attention anyway! There was a side-debate on smoothing; I personally love well-done smoothing, especially for low-abundance MS data, because integrators do a much better job when not fooled by noise, but others are quite adamant it's totally unacceptable as manipulation of the data.
Secondly, Chemstation does external and internal calibration curves quite nicely (note: you have to choose whether you want internal or external in define report, not in set up calibration!), certainly with more convenience than exporting to Excel. Simply define your signals, define your peaks in the calibration table, and add levels for all your standards. Enter into "amount" the amounts in each standard. The table will be calibrated as you go along if you open each data file for a standard, and use "add level" to add its amount. Minor note: "new calibration table" messes up some signal details (particularly smoothing) which need repairing after starting a new cali table. If you've set up a cali table once, and you are recalibrating, you can open the standard and click on recalibrate, and it will change the corresponding point in the calibration curves. These can all be viewed, exported etc.
Batch processing, assuming you have not defined samples as standards in the sequence, uses the existing calibration curves embedded in the method to calculate all samples. It works differently to running a sequence where standards are defined, in that there is no bracketing: it simply uses the embedded curve for the entire batch, with no changes.
Good luck!