I used Partisil SAX column to analyse bromide in polysulphide. It worked pretty good but, suddenly lost resolution.
Can I regenerate the column?How?
That depends on how/why it lost resolution. Resolution is a function of retention, selectivity (the relative spacing of the peaks), and efficiency (peak width). So what changed?

If your peaks are in the right place, but too wide, backflushing the column can clear a partially plugged inlet or (temporarily) undo a void space. If retention has decreased significantly, then flushing the column with a higher-ionic-strength buffer to displace strongly-bound contaminants can help.

That column is silica-based, so it can tolerate high levels of organic (methanol would be preferred) to flush off hydrophobic contamination.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 76 users online :: 0 registered, 0 hidden and 76 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 76 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry