Sulfate, Nitrate and Nitrite determinationReverse phase HPLC
Posted: Fri Aug 30, 2013 1:02 pm
by alper
Hi,
I can determine Cl, NO3 and NO2 by using C18 column with the eluent n-Octylamine at 220 nm. However, I want to measure Sulfate additionally.
My methode is below
Column Eurosil Bioselect 300-5 C18, 120 x 4 mm
Eluent A 0.01 M n-Octylamine, set to pH 4
Flow rate 1.0 ml/min
Injection volume 5 µl
Column temperature 40 °C
Detection UV at 210 nm
Do you have an application to measure sulfate nitrate and nitrite together with HPLC-UV
Thanks
Re: Sulfate, Nitrate and Nitrite determinationReverse phase
Posted: Fri Aug 30, 2013 4:39 pm
by tom jupille
What else is in your mobile phase that you neglected to mention? Cl- is pretty much transparent at UV wavelengths, so you are *not* detecting it directly at 210 nm. Sulfate has the same issue: it does not absorb UV light.
Chloride and sulfate can be done via indirect UV (for example:
http://www-user.uni-bremen.de/~koelling/so4.html) but it's much more straightforward with an ion-exchange column rather than ion-pair.
Re: Sulfate, Nitrate and Nitrite determinationReverse phase
Posted: Fri Aug 30, 2013 5:53 pm
by alper
Hi Tom. Mobile phase is 0.01 M n-Octylamine at pH 4. But Once I checked Cl. I saw peak.
Re: Sulfate, Nitrate and Nitrite determinationReverse phase
Posted: Fri Aug 30, 2013 10:46 pm
by HPLCaddict
Hi Tom. Mobile phase is 0.01 M n-Octylamine at pH 4. But Once I checked Cl. I saw peak.
Probably the solvent front?
Re: Sulfate, Nitrate and Nitrite determinationReverse phase
Posted: Sat Aug 31, 2013 12:45 am
by tom jupille
That size column would have an internal volume of around 1 mL. At 1 mL/min, that makes the dead time about 1 minute. Certainly anything eluting before 2 minutes is t0 noise. It's not good practice to quantitate any peak eluting with a k' less that 2 (in this case, retention time less than about 3 minutes).
Re: Sulfate, Nitrate and Nitrite determinationReverse phase
Posted: Sun Sep 01, 2013 12:44 am
by juddc
Ion exchange with conductivity detection is the simplest solution to this and an old conductivity detector should be easy enough to find and wouldn't cost much. I've run the same unit on and off for 20 years, have never done a thing to it, and it runs fine.
For $500, it'll be less expensive than spending half a day working on your RP method:
http://www.ebay.com/itm/Waters-431-Cond ... 3cd2df570d
Re: Sulfate, Nitrate and Nitrite determinationReverse phase
Posted: Tue Sep 03, 2013 9:41 am
by ACrawshaw
Hi Alpher,
I would have to concur with the previous comments above. Have you considered IC? Specifically designed for what you are trying to do...
http://www.dionex.com/en-us/products/io ... 72590.html
Thanks
Re: Sulfate, Nitrate and Nitrite determinationReverse phase
Posted: Wed Sep 04, 2013 2:28 pm
by Markus Laeubli, Metrohm
To add up an other IC player:
View
Metrohm's IC
Re: Sulfate, Nitrate and Nitrite determinationReverse phase
Posted: Wed Sep 04, 2013 3:39 pm
by DJ
I've seen this done with a C18 "permanently coated" with some quaternary ammonium modifier with hydrophobic arms. The column never lasts long.
Friend, you could probably separate sulfate, nitrite, nitrate on a SAX guard column. Why re-invent wheel?
Re: Sulfate, Nitrate and Nitrite determinationReverse phase
Posted: Thu Sep 05, 2013 10:08 pm
by Vlad Orlovsky