If you have resolved your mix, on the first step is identify each peak. Open you chromatogram, go to method section, go to integration events and obtain the integrations for the each peak of your interest. After that, go to peak identification and rigth click . Choice "initialize from chromatogram", put the names for each peak. and integrate F5, after this, on the chromatogram are the names. Now go to calibration section and define the calibration type, levels of calibration, units, give a name for your calibration file. Click on initialize from ID tables. Now put the concentraction value for each peak and level. Go to save, and click on the save "cromato method". This, save the changes on the method analysis. Close your method analysis, if it is open, this refresh the changes on your method.
Now open your chromatograms of each level calibration, go to reproces an click, in the new window, on the chromatogram, select the first chromatogram, normally the lowest concentratation, select your method analysis wich contain your calibration file. On the sample type select "standard", on calibration mode select "clear old points", now select calibration level, normally 1. Repeat all of this with the others chromatograms, but for all of them select in calibration mode "add". Close your reprocess window.
If all is OK, go to calibration section, on the bottom of your main page and open your calibration file. You can observe the calibrations curves. At this moment, your method analysis is now a cuantitation methos, when you analyze samples you can obtain the concentratio of each compound.
I hope yhis help.
E.M. Calidad AnalĂtica S.A. de C.V.
emcasa_cv@hotmail.com