varian 3800 GC
Posted: Wed Aug 28, 2013 5:10 am
by ramesh61
Hi I have varian 3800 GC. when I am not using Capillary Injector (capillary Column is not attched) I still get 12.5 ml flow in 1177 EFC?
I am geting not ready message and hence not able to run the GC?What clule be Peoblem?
Re: varian 3800 GC
Posted: Thu Aug 29, 2013 7:55 am
by Peter Apps
Ifr you are not using that inlet then set the flow and pressure to zero and plug the column connection with a nut and blank ferrule.
Re: varian 3800 GC
Posted: Fri Aug 30, 2013 4:33 pm
by danicrd
Have you checked setup of instr? in column section, have you left original column configured? For example 30m *.025?
sometimes 3800 gc require fake configuration on inlet, even if you don't install col: so enter setup, and insert 1m*0.25 in conf and reboot
Sometimes with my 3800 has worked
Re: varian 3800 GC
Posted: Sat Nov 30, 2013 1:05 pm
by ramesh61
:lol:
Thnks for replay
I have done that
Now change of topic
In other unit of varian 3800 GC i am unable to make injection beacuse Gc is nit geting ready due to flow not geting stabilised by EFC. I checked al connection for leakages,The Septa changed,Capillary column leakage cheked
The Pressure is 10 PSI and I get set pressure 10 psi and actual pressure is 45 PSI ? What is the problem?