hiii dear all.I am using API 2000 LC-MS/MS System for Impurity profiling purpose. While using this,it was found that in mass spectra a molecular ion (M+H+38) larger than parent compound (M) always. What is that ?? Pleazzzzzz guide me also suggest to avoid that >>peaks
Potassium adduct ?

Try increasing the declustering voltage.
-- Tom Jupille
LC Resources / Separation Science Associates