mess up with gas tank for GC/MS
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
11 posts
Page 1 of 1
used nitrogen instead of helium. Could it cause any problem? just run like twenty minutes.
-
- Posts: 1680
- Joined: Sat Aug 23, 2008 12:04 am
It depends. If the mass spec was off - not a problem.
If the mass spec was on - maybe. If you were acquiring m/z 28, you may have beat up your detector a bit. With the instrument switched over to He and the lines carefully purged, tune the instrument and see how it does. If you get a good tune with about the same signal intensity at the about the same detector voltage - no harm done.
If the mass spec was on - maybe. If you were acquiring m/z 28, you may have beat up your detector a bit. With the instrument switched over to He and the lines carefully purged, tune the instrument and see how it does. If you get a good tune with about the same signal intensity at the about the same detector voltage - no harm done.
-
- Posts: 418
- Joined: Thu Sep 25, 2008 3:40 am
Normally, the software will prompt an error message because you have to define the carrier gas as He when editing the method.
-
- Posts: 1680
- Joined: Sat Aug 23, 2008 12:04 am
Ths sounds like putting a nitrogen bottle on the regulator where it should have been a helium bottle - and they can be switched this way. The software has no idea what is actually coming through the plumbing.
-
- Posts: 693
- Joined: Mon Mar 28, 2005 5:38 pm
Agilent is now selling a device for the 7890 GCs to switch your carrier gas to nitrogen while at idle. For those of us with instruments that are frequently parked it could save quite a bit of helium. Of course those instruments don't get replaced very often so maybe by the time I retire...
-
- Posts: 418
- Joined: Thu Sep 25, 2008 3:40 am
The software does not know what gas it is, but the gas viscosity is different so it fails to maintain the right flowrate.Ths sounds like putting a nitrogen bottle on the regulator where it should have been a helium bottle - and they can be switched this way. The software has no idea what is actually coming through the plumbing.
-
- Posts: 206
- Joined: Sun Oct 09, 2011 5:08 pm
Thanks god. I did not acquire 28, nitrogen gas. so that is fine.
It depends. If the mass spec was off - not a problem.
If the mass spec was on - maybe. If you were acquiring m/z 28, you may have beat up your detector a bit. With the instrument switched over to He and the lines carefully purged, tune the instrument and see how it does. If you get a good tune with about the same signal intensity at the about the same detector voltage - no harm done.
-
- Posts: 1113
- Joined: Wed Dec 22, 2010 10:17 pm
Other than possibly some wear on the filament it shouldn't be an issue.
-
- Posts: 377
- Joined: Wed Aug 24, 2011 2:10 pm
A valve? really that's all it would take, and just set the gas to nitrogen in yoru standby method, helium in your analysis method. Chemstation would probably flip out at this point and report some mismatch, but Openlab doesn't, it just gets on with it (see, there are some advatages!). I guess there will always be people who can't turn a valve though, and Agilent have answer for them!Agilent is now selling a device for the 7890 GCs to switch your carrier gas to nitrogen while at idle.
Where can I buy the kit they use in CSI?
-
- Posts: 5432
- Joined: Thu Oct 13, 2005 2:29 pm
The trouble is that the accreditation cops would require the valve to be certified, there would have to be an SOP for when and how to turn the valve, and everyone would have to have training in valve turning, suitably documented in their training files. Then each time the valve was turned it would have to be documented and signed off. So it's actually much easier to include it in the instrument setup and let the software do the paperwork. Sad really.A valve? really that's all it would take, and just set the gas to nitrogen in yoru standby method, helium in your analysis method. Chemstation would probably flip out at this point and report some mismatch, but Openlab doesn't, it just gets on with it (see, there are some advatages!). I guess there will always be people who can't turn a valve though, and Agilent have answer for them!Agilent is now selling a device for the 7890 GCs to switch your carrier gas to nitrogen while at idle.
Peter
Peter Apps
-
- Posts: 3477
- Joined: Mon Jan 07, 2013 8:54 pm
Yup, takes more lawyers than scientist to run a sample these days.The trouble is that the accreditation cops would require the valve to be certified, there would have to be an SOP for when and how to turn the valve, and everyone would have to have training in valve turning, suitably documented in their training files. Then each time the valve was turned it would have to be documented and signed off. So it's actually much easier to include it in the instrument setup and let the software do the paperwork. Sad really.A valve? really that's all it would take, and just set the gas to nitrogen in yoru standby method, helium in your analysis method. Chemstation would probably flip out at this point and report some mismatch, but Openlab doesn't, it just gets on with it (see, there are some advatages!). I guess there will always be people who can't turn a valve though, and Agilent have answer for them!Agilent is now selling a device for the 7890 GCs to switch your carrier gas to nitrogen while at idle.
Peter
The past is there to guide us into the future, not to dwell in.
11 posts
Page 1 of 1
Who is online
In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google Adsense [Bot] and 1 guest
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google Adsense [Bot] and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science