- Posts: 9
- Joined: Tue Aug 07, 2012 3:58 pm
Equipment:
2 x SD-200
1 x FC-1
1 x UV-1
1 x Original iMac ("Bondi Blue") running Rainin Dynamax Method Manager 1.4
1 x Keyspan USA-28XG USB->Serial adapter
The hardware all works (well, as of 1.5 years ago when it was turned off and not touched) but I'm having trouble getting the serial communications going between the computer and the instrument(s).
DMM apps run and captures the "modem" port (port 2 on the USA-28XG). I can verify that traffic is flowing in both directions across the USB/serial interface (the counters increment). In the DA "Device Status" window, I get "Err: -16990" for all ID's that correspond with attached devices. I configured the 9-pin ribbon cable (Pump A is at the end, B is second from end, on the third connector) and assigned ID's per the manual.
If I disconnect a device, it shows as offline. Reconnecting the device gives me the -16990 error again.
I've connected just one device at a time and the results are the same.
I'd appreciate any advice that anyone can give.