one compound has too much fragmentation, so sensitivity is very low. How can I decrease the fragmentation and increase the sensitivity.
Chromatography Forum is a public discussion group where you can post questions, news, or messages of interest to chromatographers everywhere, but you must be registered to participate (registration is FREE). If you are a registered user, please log in.
http://www.chromforum.org/
Extract an ion of interest and use that, or use SIM (selected ion monitoring).one compound has too much fragmentation, so sensitivity is very low. How can I decrease the fragmentation and increase the sensitivity.
Extract an ion of interest and use that, or use SIM (selected ion monitoring).one compound has too much fragmentation, so sensitivity is very low. How can I decrease the fragmentation and increase the sensitivity.
One option on an Agilent 5973 or 5975 would be to lower the Emission setting which decreases the current (but not the Electron Volts) at the filament. You will sacrifice a little overall sensitivity but it will bias the fragmentation towards higher masses because of less total fragmentation. The ion fragments should be the same, just biased more to the heavier masses.
If it is a compound with multiple chlorine molecules, well there isn't much you can do in that case, all the permutations of having multiple chlorine molecules with 35 and 37 masses causes a lot of fragments and lowers the intensity of each.