Measurement of Tween 80 in Pharmaceuticals
Posted: Wed Aug 03, 2005 3:31 pm
by KarenJ
Does anyone know of a method for quantitating the levels of Tween 80 (polysorbate 80) in solution formulations? I've seen one reference that uses Evaporative Light Scattering detection. The boiling point of Tween 80 is listed in the MSDS as 100C. Will ELSD work well for an analyte with a boiling point this low?
Karen J
Posted: Wed Aug 03, 2005 4:24 pm
by Mark Tracy
Tween-80 has various amounts of water in it, hence the low boiling point. Actually, it works well on an evaporative light scattering detector. The bad news is that it is a complex mixture with many peaks. We have run it here on our Acclaim Surfactant column, but I don't have a link to a sample chromatogram.
Quantitation Tween 80
Posted: Wed Aug 03, 2005 5:53 pm
by JMB
Karen,
L. M. Nair et al. Journal of Chromatography A
Volume 1012, Issue 1 , 12 September 2003, Pages 81-86
This is probably the ref. that you mention. Chromatography to give single peak, with ELSD is used.
LOQ = 40 ug/mL and LOD = 15 ug/mL are quoted.
JMB
tween 80 assay
Posted: Thu Aug 04, 2005 9:30 pm
by tom_mizukami
Tween 80 is a complex mixture, I think I would stay away from HPLC.
You could measure the uptake of a fluorscent dye in or onto tween micelles. Here is a paper
http://wwwsoc.nii.ac.jp/jsac/analsci/pdfs/a19_0737.pdf
Of course you could use an HPLC with a restriction capillary or reaction coil instead of a column.
Tom
Posted: Thu Aug 04, 2005 10:02 pm
by JMB
Tom,
Yes, Tween 80 certainly is a complex mixture, however the reference that I cited describes the use of a very steep gradient (MeOH 30 ---> 90 %, 4 min) on a C-18 column to compress the many peaks into a single (fronting) peak for quantitation purposes.
JMB
Posted: Fri Aug 05, 2005 8:52 pm
by tom_mizukami
Hi JMB,
I'm not doubting that it can be done, but why bother with a column at that point.
Tom
Posted: Sun Aug 07, 2005 6:06 am
by JM
Why you need to quantify a excepient , which is mix of several components?? normally this type of excepient ( like Tween-80 and Cremophor ) not quntified but a id test is done . If you try to elute all of the tween-80 componets by high organic chances are that all your drug product will also elute. Also pls note this type of natural originated exceipient components vary with the time and region .
JM
Posted: Mon Aug 08, 2005 12:58 pm
by KarenJ
JM,
We aren't qualifiying the Tween 80 as an excipient. We have been asked to compare the levels of Tween in several finished parenteral products.
Karen J
Tween 80 Quantn
Posted: Mon Aug 08, 2005 7:37 pm
by JMB
Tom--I think Karen needs to measure Tween 80 in formulations, so cannot use FIA since ELSD is not selective.
JM--I take the point about drug also eluting at high %ORG, however the paper shows a satisfactory example. Their description of the "gradient" is a little ambiguous--seems probable that they run at 30%ORG for 4 min then step to 90%ORG to drive off the Tween as a single peak while leaving the drug on the column (rather than a linear 30 --> 90% org ramp).
Karen-- IF you try the method and IF it works for your API/Tween formulation please let us know.
JMB
Measuring Tween 80 in Pharmaceutical Formulations
Posted: Mon Aug 08, 2005 8:12 pm
by KarenJ
Thanks to everyone for their input. I will let you know if we end up trying this.
KarenJ