Plasticisers in biological matrices
Posted: Tue Aug 20, 2013 3:29 pm
by steve wileman
I'm detecting plasticisers (mainly phthalate metabolites and BPA) in biological matrices using SPE - LC-MS/MS. I have been having trouble getting my QC's to pass. ie I can produce pretty nice calibration plots in one batch of urine (for example), but when I spike a different batch I get a different response - particularly at the lower end (1-5ng/ml). I am measuring what is in the matrix to start with (ie the "blank") and deducting this from the total peak area. I use C18 SPE, then either phenyl or C18 HPLC column.
Has anybody else come across this difficulty of getting "blank" biological matrices?
Thanks!
Steve
Re: Plasticisers in biological matrices
Posted: Wed Aug 21, 2013 2:10 am
by Don_Hilton
These compounds are quite common in the general population
http://www.cdc.gov/exposurereport/pdf/FourthReport.pdf It may be worth checking the procedures used by laboratories such as the CDC to see where they get materials to use as "Blank" matrix. A look at the published papers may help - or look for authors and contact them. And the better your analytical method - the harder it is going to be to find "blank" urine.
Posted: Wed Aug 21, 2013 2:47 pm
by Steve Reimer
Or even "blank" containers to collect and process it in. We have had problems getting fish that is free of mercury for low level mercury analysis.
Posted: Fri Aug 23, 2013 2:21 pm
by steve wileman
Yes, thanks for that. I recognise what you're saying - I have tried spiking artificial urine which works well, but I dont get a true picture when compared to the real matrix ie. the peaks are much bigger at a fixed concentration. Others have also produced cali-plots in solvent, which is not representative of the matrix at all. I can deal with non-blank blanks provided I measure what that is, and deduct it - the problem seems to be that when I attempt to replicate cali-plots in a different batch of matrix, and even after deducting the blank values for this new matrix, the concentrations work out different - especially at low concentrations - 1-5ng/ml. Having said all that, I have been experimenting with a 1mm HPLC column (as opposed to 2.1mm) and at least for some analytes the situation has improved. This may indicate that it is an integration/sensitivity problem - always tricky when working at these trace levels! Thanks again for your response.
Posted: Tue Sep 03, 2013 6:42 am
by Alp
One possibility is that something in some urine batches (but not others) is co-eluting with your analyte and causing matrix effect. Are you retaining your analyte on an HPLC column and if so, is it retaining well outside the void volume?
Posted: Wed Sep 18, 2013 11:34 am
by steve wileman
thanks very much for that, I also think that this is possible, but analytes are retained until well after void volume I would say. Things are looking a bit better with the 1mm columns and also the Phenomenex core shell column. Nearly there.. Thanks again.