It seems I chose the wrong reference wavelength. So my peaks are negative. Does anyone know if I can change the reference wavelength and reprocess the data or can I use "isoabsorbace plot" function or other function to somehow correct it?
-
- Posts: 39
- Joined: Wed Jun 15, 2005 2:53 am
Yes you can extract at any wavelength and with any refernce wavelength. But the data should be acquired with all data points and in complete uv range. Can you please describe the complete signal details for acquiring data.
-
- Posts: 49
- Joined: Tue Nov 23, 2004 1:31 pm
Dear pawan ratra,
I work with Chemstation but not a good detector because it is a variable wavelength detector. I do not know which is your detector but I think in the worst situation that could a fixed wavelength detector. You coul work with a standard solution and a spectrophotometer where you could see the spectrum and select the best wavelength.
Good luck,
Diego Delmonte
I work with Chemstation but not a good detector because it is a variable wavelength detector. I do not know which is your detector but I think in the worst situation that could a fixed wavelength detector. You coul work with a standard solution and a spectrophotometer where you could see the spectrum and select the best wavelength.
Good luck,
Diego Delmonte
-
- Posts: 29
- Joined: Fri Nov 12, 2004 10:49 am
Dear Apple_insci
If You work a reference wavelenght in ChemStation, so you use a DAD detector, isn't? So you can determine the DAD spectra of your peaks, and the peak of your analyte. It is easy, but if you have any problem... contact the Agilent Local Office in you country or send us notice.
Have good elutions!!!
Carlos Teixeira
If You work a reference wavelenght in ChemStation, so you use a DAD detector, isn't? So you can determine the DAD spectra of your peaks, and the peak of your analyte. It is easy, but if you have any problem... contact the Agilent Local Office in you country or send us notice.
Have good elutions!!!
Carlos Teixeira
4 posts
Page 1 of 1
Who is online
In total there are 25 users online :: 0 registered, 0 hidden and 25 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 25 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 25 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science