Thank you danny and david,
I also did the cal gas adjustment and made the cal gas pressure to the middle range. Any how I will try to arrange a ion trap cleaning
Below are the full details I got from the auto tunning. Hope this will help you to get an idea about the problems. Again thanks soo much for the great help ..
10:05:04 Auto Tune: Started
10:05:05 Air/Water Check: Started
10:05:31 Air Check: Acceptable Level Found (28 Width: 0.6 m/z)
10:05:32 Water Check: Acceptable Level Found (19/18 Ratio: 8.5 %)
10:05:32 Air/Water Check: Completed - No Problems Found
10:05:33 Integrator Zero Set: Started
10:05:41 Integrator Zero Set: Setting is OK (Setting: 130, Average Counts: 0.50)
10:05:42 Integrator Zero Set: Completed
10:05:42 Electron Multiplier: Started
10:05:44 Electron Multiplier: Offset Determined (Offset: 1 Count(s))
10:05:59 Electron Multiplier: Pre-Adjustment Successful (EM Voltage: 1400)
10:09:54 Electron Multiplier: Low Voltage End Found (EM Voltage: 1300)
10:11:03 Electron Multiplier: High Voltage Start Found (EM Voltage: 1500)
10:11:12 Electron Multiplier: Space Charge Adjusted (Target: 20000, Val/Iso: 0.25)
10:11:52 Electron Multiplier: Peak Threshold Not Met (Inten: 168 counts)
10:11:53 Electron Multiplier: 10^5 Gain setting is OK (EM Voltage: 1400)
10:11:53 Electron Multiplier: Final Gain setting is Marginal (EM Voltage: 1700)
10:11:54 Electron Multiplier: Completed
10:11:54 RF Full Scale Adj: Started
10:12:02 RF Full Scale Adj: Centered on Mass 69 at 69.91 (Setting: 60)
10:12:07 RF Full Scale Adj: Centered on Mass 414 at 416.19 (Setting: 33)
10:12:11 RF Full Scale Adj: Centered on Mass 614 at 642.55 (Setting: 1)
10:12:12 RF Full Scale Adj: Setting is High (Setting: 1, Mass: 614, Apex: 643.

10:12:12 RF Full Scale Adj: Completed
10:12:13 Multi-Point Mass Cal: Started
10:12:19 Multi-Point Mass Cal: Found Calibration Mass 28 at 27.63
10:12:26 Multi-Point Mass Cal: Found Calibration Mass 69 at 68.25
10:12:31 Multi-Point Mass Cal: Found Calibration Mass 131 at 129.79
10:12:42 Multi-Point Mass Cal: Found Calibration Mass 264 at 262.07
10:13:44 Multi-Point Mass Cal: Found Calibration Mass 502 at 499.03
10:14:11 Multi-Point Mass Cal: Found Calibration Mass 614 at 622.54
10:14:12 Multi-Point Mass Cal: Calibration is OK (Slope: 6.243, Std Dev: 0.063)
10:14:12 Multi-Point Mass Cal: Completed
10:14:13 Trap Function Calib: Started
10:14:16 Trap Function Calib: Calibrating at Mass 69
10:15:21 Trap Function Calib: Calibrating at Mass 131
10:15:22 Trap Function Calib: No Reference Peak Located
10:15:22 Trap Function Calib: Failed
10:15:23 Auto Tune: Terminated
10:28:20 Auto Tune: Started
10:28:21 RF Full Scale Adj: Started
10:28:30 RF Full Scale Adj: Centered on Mass 69 at 69.92 (Setting: 59)
10:28:35 RF Full Scale Adj: Centered on Mass 414 at 416.05 (Setting: 34)
10:28:39 RF Full Scale Adj: Centered on Mass 614 at 643.67 (Setting: 1)
10:28:40 RF Full Scale Adj: Setting is High (Setting: 1, Mass: 614, Apex: 644.0)
10:28:40 RF Full Scale Adj: Completed
10:28:41 Multi-Point Mass Cal: Started
10:28:47 Multi-Point Mass Cal: Found Calibration Mass 28 at 27.63
10:28:55 Multi-Point Mass Cal: Found Calibration Mass 69 at 68.25
10:29:00 Multi-Point Mass Cal: Found Calibration Mass 131 at 129.79
10:29:11 Multi-Point Mass Cal: Found Calibration Mass 264 at 262.08
10:29:30 Multi-Point Mass Cal: Found Calibration Mass 414 at 411.42
10:30:13 Multi-Point Mass Cal: Found Calibration Mass 502 at 498.98
10:30:40 Multi-Point Mass Cal: Found Calibration Mass 614 at 622.62
10:30:41 Multi-Point Mass Cal: Calibration is OK (Slope: 6.243, Std Dev: 0.063)
10:30:41 Multi-Point Mass Cal: Completed
10:30:42 Auto Tune: Completed
10:31:41 Auto Tune: Started
10:31:42 Trap Function Calib: Started
10:31:47 Trap Function Calib: Calibrating at Mass 69
10:32:44 Trap Function Calib: Calibrating at Mass 131
10:33:46 Trap Function Calib: Mass 69 Calibrated (Freq: 262.800 kHz)
10:33:46 Trap Function Calib: Mass 131 Calibrated (Freq: 259.900 kHz)
10:33:47 Trap Function Calib: Completed
10:33:47 Auto Tune: Completed