Hi Guys

I am currently developing an ion chromatography method for the simultaneous analysis of Cl, Br and I. This is my first time performing a proper validation and I have a question regarding calibration.

From the guidelines I have been following to perform my validation, it states that 'the blank matrix sample should not be taken into consideration when plotting calibration curves'. I was wondering wether this approach was taken by others who have previous experience of method validation? How do you construct your graph i.e. force point through zero? plot blank sample? only plot validation standards minus the blank sample?

Thank you for your help in advance.
force point through zero?
No
plot blank sample?
No
only plot validation standards minus the blank sample?
Yes
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Hi Tom

Thank you for your reply - it is much appreciated.

Could I please have some clarification on the 'only plot validation standards minus the blank sample' statement I made? I dont think I worded it very well.

Does this mean:

(A) I only plot the calibration standards
(B) I only plot the calibration standards (after subtracting any blank matrix areas from the calibration standards areas)

Thank You
(A) I only plot the calibration standards
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
I would strongly suggest that your calibration standards match the matrix of your samples. Use a clean sample to make calibration standards and also inject it as a matrix blank.

I don't know what your application is but from my experience with chromatography, a calibration curve should always be made using matrix matched standards.
BHolmes

Any problem worthy of attack, proves its worth by hitting back...never give up!
a calibration curve should always be made using matrix matched standards.
I agree, but that wasn't the question. The question was interpretation of the statement 'the blank matrix sample should not be taken into consideration when plotting calibration curves'
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
For what it's worth, we also don't have a calibration blank in our curve. Some methods require a force through zero but other than that we don't. Only the standards themselves are included. The calibration black can be read as a qc standard though.
7 posts Page 1 of 1

Who is online

In total there are 49 users online :: 0 registered, 0 hidden and 49 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 49 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry