Running robustness tests
Posted: Thu Aug 15, 2013 5:58 pm
I am trying to determine the more preferred approach to testing the robustness parameters during a method validation. I want to limit the scope to instrument-controlled parameters (e.g. flow rate, temperature, wavelength, mobile phase composition).
Do I run each particular adjustment as a separate sample set (standards and samples run at the same condition) or can I run the changes in a single sample set, using the procedure-stated instrument parameters for my standard injections and samples as my baseline, then adjust the component parameters using different instrument methods, comparing the sample results with my baseline samples?
On a related topic, when a DOE is used to examine robustness, are the standards run at the procedure-stated parameters and then the DOE parameters are changed for the samples, with the samples quantitated from the "nominal" conditions?
Do I run each particular adjustment as a separate sample set (standards and samples run at the same condition) or can I run the changes in a single sample set, using the procedure-stated instrument parameters for my standard injections and samples as my baseline, then adjust the component parameters using different instrument methods, comparing the sample results with my baseline samples?
On a related topic, when a DOE is used to examine robustness, are the standards run at the procedure-stated parameters and then the DOE parameters are changed for the samples, with the samples quantitated from the "nominal" conditions?