This may be a silly question, but it has been confusing me.

I am making some injections of known concentrations for my experiments. For some of the calculations I want to make, I need the concentration of my analytes in the mobile phase to produce the correct graph.

Do I need to solve some differential equation (since conc. is a function of time and position in the column)? Or could I just use the void volume of the column as a factor that is diluting my known injection concentration?

Thanks for any assistance!
The time when your Peak starts and Ends will give you the volume in which your compound is solved. If flow is 1ml per min and this time is half a Minute your analyt is disolved in half a ml. Maybe that Explanation is too easy and some colleagues will give you the right equation. We will see.
Gerhard Kratz, Kratz_Gerhard@web.de
The time when your Peak starts and Ends will give you the volume in which your compound is solved. If flow is 1ml per min and this time is half a Minute your analyt is disolved in half a ml. Maybe that Explanation is too easy and some colleagues will give you the right equation. We will see.
That seems to makes sense. I will try that out with some of my calculations and see if the results make sense. Thank you.
The time when your Peak starts and Ends will give you the volume in which your compound is solved. If flow is 1ml per min and this time is half a Minute your analyte is dissolved in half a ml.
That assumes that your starting point was the peak area (i.e., that you have already integrated the peak). If you are using peak height, and you have a Gaussian peak, you need to divide by 2 (the maximum amplitude is approximately twice the average).
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
4 posts Page 1 of 1

Who is online

In total there are 56 users online :: 2 registered, 0 hidden and 54 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Semrush [Bot] and 54 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry