Riboflavin ISO Guide 34
Posted: Fri Aug 09, 2013 8:56 pm
by riterkl
We have recently purchased a riboflavin (vitamin B2) analytical standard which conforms to ISO Guide 34 and are having trouble dissolving normally in warm dilute acetic acid as in AOAC 970.65. Has anyone else experienced this and how were you able to dissolve the standard? Any help is greatly appreciated. Thanks.
Re: Riboflavin ISO Guide 34
Posted: Fri Aug 09, 2013 10:12 pm
by Don Shelly
The method calls for 50 mg B2 in 300 ml of heated acetic acid with frequent agitation. Keep heating until the standard goes into solution. Is this the procedure that you are following?
Posted: Fri Aug 16, 2013 4:23 am
by Meerkat
Also, you may want to try ultra-sonication; however, be sure to use cold water in the sonicator bath. Put the vol flask containing your solvent and powdered vitamin in the sonicator with cold water bath and sonicate for 5 to 10 mins with regular shaking. I've used this with cyanocobalamin vit b12, which was also difficult to solvate.
Posted: Sat Aug 24, 2013 2:32 am
by riterkl
Yes, this is the procedure we are following, but it does not want to dissolve.
Posted: Sat Aug 24, 2013 2:33 am
by riterkl
I think that we have tried ultrasonic bath, but I will double-check this. Planning an experiment on Monday, so maybe I will have some news then.
Posted: Sat Aug 24, 2013 2:36 am
by riterkl
Maybe one more note is that we can dissolve USP riboflavin standard with no problems for many years. Only since we have gone ISO 17025 did we try guide 34 standard.