HPLC post-calculations...
Posted: Thu Aug 08, 2013 9:21 am
HI all.
I could really use some help here. I´m currently working on a method for measuring creatinine in serum from mice, and have decided to use HPLC for this.
I´m using the Zorbax SCX300 2.1 x 50 mm, 5 micron, with in-front guard column, and a 5 mM sodium acetate pH 4.1 (adjusted with glacial acetic acid) as mobile phase.
My standard curve is composed of 9 concentrations of creatinine powder from Sigma (128 - 64 - 32 - 16 - 8 - 4 - 2 - 1 - 0.5 uM) dissolved in mobile phase.
This is my prep of std. and samples:
25 ul std./sample
100 ul cold Acetonitrile (AcN) + 0.5% glacial acetic acid to precipitate proteins
vortex 15 s
15 min at -20 degrees
spin 10000 rpm at 4 degrees for 10 min
transfer sup. to new epp. (100 ul of the initial 125 ul ~ 80%)
speedvav to (~ 40 -45 min)
resuspend in 25 ul mobile phase
spin 10000 rpm at 4 degrees for 10 min
transfer all 25 ul to 96-well Abgene PCR plate for Aqurity autosampler
load 3 ul/run (all samples in duplicates)
run time: 10 min
flow: 1 ml/min
column temp: 50 degrees
autosampler temp: 18 degrees
backpressure: ~ 2300 psi
UV at 225 nm
My standard curves are pretty similar with an R square of 0.96-0.99
But, and this is my real question, when I interpolate my samples with an unknown concentration of creatinine how do I calculate the final concentration of creatinine in the initial serum sample??? In either umol/L or mg/dL??
I could really use some help here. I´m currently working on a method for measuring creatinine in serum from mice, and have decided to use HPLC for this.
I´m using the Zorbax SCX300 2.1 x 50 mm, 5 micron, with in-front guard column, and a 5 mM sodium acetate pH 4.1 (adjusted with glacial acetic acid) as mobile phase.
My standard curve is composed of 9 concentrations of creatinine powder from Sigma (128 - 64 - 32 - 16 - 8 - 4 - 2 - 1 - 0.5 uM) dissolved in mobile phase.
This is my prep of std. and samples:
25 ul std./sample
100 ul cold Acetonitrile (AcN) + 0.5% glacial acetic acid to precipitate proteins
vortex 15 s
15 min at -20 degrees
spin 10000 rpm at 4 degrees for 10 min
transfer sup. to new epp. (100 ul of the initial 125 ul ~ 80%)
speedvav to (~ 40 -45 min)
resuspend in 25 ul mobile phase
spin 10000 rpm at 4 degrees for 10 min
transfer all 25 ul to 96-well Abgene PCR plate for Aqurity autosampler
load 3 ul/run (all samples in duplicates)
run time: 10 min
flow: 1 ml/min
column temp: 50 degrees
autosampler temp: 18 degrees
backpressure: ~ 2300 psi
UV at 225 nm
My standard curves are pretty similar with an R square of 0.96-0.99
But, and this is my real question, when I interpolate my samples with an unknown concentration of creatinine how do I calculate the final concentration of creatinine in the initial serum sample??? In either umol/L or mg/dL??