which fragment ion should be used for quantification

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
generally base peak, or the second largest peak?
I usually choose one that is unique and hopefully quite larger. A unique fragment makes most of the other stuff you don't care about vanish and a large one gets you the sensitivity you want. You should also monitor some others that are characteristic to your analyte so that you don't identify a coeluter as your analyte. The ratios of the fragments should be characteristic of your analyte. Generally, even isomers have different ratios of the fragments generated.

Good luck!
make sense. thanks.
I usually choose one that is unique and hopefully quite larger. A unique fragment makes most of the other stuff you don't care about vanish and a large one gets you the sensitivity you want. You should also monitor some others that are characteristic to your analyte so that you don't identify a coeluter as your analyte. The ratios of the fragments should be characteristic of your analyte. Generally, even isomers have different ratios of the fragments generated.

Good luck!
3 posts Page 1 of 1

Who is online

In total there are 23 users online :: 0 registered, 0 hidden and 23 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 23 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry