calibration of mass spectrometer

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
Hi guys,

I would like to know something about instrument calibration. (Although I do not need to take charge of that =P)
The MS instrument is Waters ACQUITY TQD, NaI/CsI solution is commonly used as calibration solution.
And I would like to know if it is possible for those ions to attach the quadrupole. Are there any reason why NaI/CsI is good for calibration?
I am just curious and would be grateful if you reply. Thanks =]
NaCsI won't stick to the quad, but they will foul a bit the source. I think that now Waters suggests to use sodium formate for calibration (that's what I use with success)

(from the site http://masstips.blogspot.be/)

Recipe for Sodium salts used to calibrate TOF
For Sodium formate:

10ml Water
10ml Isopropanol
40ul Formic Acid
200ul NAOH (1 molar)
All solvents must be of highest quality.
2 posts Page 1 of 1

Who is online

In total there are 168 users online :: 1 registered, 0 hidden and 167 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Bing [Bot] and 167 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry