My problem:
Both my ASRS ultra II and CSRS Ultra II have been leaking intermittently from the seam. Dionex tells me that there has the be a downstream pressure problem but I can't find one.

My setup:
I'm running a DX-600 system with an EG-40 Eluent Generator (KOH). The EG-40 has a CR-ATC in it. It is a 4mm peek system with a std bore pump.
Downstream from the ASRS eluent out port looks like this:
ASRS Eluent Out -> Column Switching Valve
CS Valve -> Cell (Conductivity)
Cell -> CS Valve
CS valve -> ASRS Regen IN
ASRS Regen Out -> CR-ATC Regen In
CR-ATC Regen Out -> EG-40 Degas Regen In
EG-40 Degas Regen Out -> Waste
Observations and Pressures Measured:(@1.0mL/min)
ASRS Eluent Out -> Waste = 66psi
Cell In -> Waste = 53psi
Cell Out - Waste = 26psi
ASRS Regen In -> Waste = 13psi
Pump on by itself = 0psi
From these pressures I concluded that the CS valve and lines create about 13psi each pass, the cell creates about 28psi. I have measure the cell before at another time and found that it created 35-40psi but there haven't been any strange things in my chromatograms.
My standards and validation samples are all within a fairly reasonable range and I don't notice any unusual background noise.
Is it possible to have the backpressure of a specific component fluctuate (maybe CS Valve) to put me over my ASRS downstream limit or is 66psi high enough to get intermittent leaks?
Thanks for any input you can provide. I appreciate it