ChemStation for Agilent
Posted: Tue Jul 30, 2013 7:29 pm
by TacioLima
I'm with difficult to generate some results with the ChemStation software Rev. B 04.03.
I'm developing a liquid chromatography method to drug quantification and, for this, i need to generate calibration curve.
But, I'm not getting put area ratio of the drug and the internal standard in Y (shows only the area of drug). Additionally not appear all my triplicates (shows just a mean in one point in the curve).
Another problem is that I cannot export the report values ​​to excel.
If someone can help me I will be very grateful.
I apologize for my english
Graciously,
Tácio
Re: ChemStation for Agilent
Posted: Wed Jul 31, 2013 3:05 pm
by MSCHemist
When you setup a calibration table in Chemstation in MSD, I'm assuming the setup is similar for Chemstation LC, you select the ITSD first and indicate it is an internal standard with either a check mark or drop down menu etc. Then you enter in the stuff for the analytes under/after the Internal standard. If your ITSD is not being integrated check your integration parameters your threshold or area reject may be set too high.
Posted: Mon Aug 05, 2013 9:42 am
by lmh
LC Chemstation is a little different to GC (although the underlying idea is very much the same). You need to go, in the data-processing review, to the menu "Report" at the top of the screen, and "Specify Report"; this opens a window, at the top right of which is the method for calculating results: by external standard, internal standard, area percent etc. You need to select ISTD.
After you've done that, you should be able to open your calibration table, and in the "Overview" view (and probably in others too) their is a column marked "ISTD" in which you can enter "Yes" for all peaks that are internal standards. For all other peaks, you can enter the number of the internal standard that you wish to use with them, and I think it goes in the column just to the right of ISTD.
One difficulty is that when you create a new calibration table, Chemstation defaults to assuming all exactly coeluting peaks are signals related to the same compound, so if you use it with mass spec and isotopically labelled internal standards, it will assume that the signal from the ISTD and target compound belong together (one will have a number in its "#" column of the cali table, and the other will be blank). To get round this, open the cali table and use "Insert" to add the 2nd compound manually.
Exporting everything to excel is very, very easy (once you know how!). If you open your sequence in the batch mode (this means looking for a file ending .b in the folder created when you ran the sequence; the file will have the same name as your sequence.s file, but will be sequence.b), then you can reprocess the sequence using whatever method you like (when you open the batch file it will offer you a chance to open it with another method, and it will ask you what files you want to process). You can set up the batch output options to include output to excel, or as csv, or whatever, and it's all completely automated. The disadvantage of post-processing is that all files are processed using the current calibration, so if you want to do bracketing, it's manual). If you need more detailed help, do ask. One warning: if you reprocess data using a different method, never ever save the batch file (or at least give it a new name). If you do so, you will never be offered the chance to choose a new method when you reopen it.
Alternatively, instead of using the built-in, reliable, easy and quick features of Chemstation, you can spend ages writing a macro, or trust someone else's dubious macro. It's amazing how many people buy software and then rewrite it instead of learning how it works.
Posted: Mon Aug 05, 2013 1:35 pm
by GasMan
If I understand the original posting requires to see the ratio of the area of the drug to the area of the internal standard as one axis of the calibration graph, you need to choose ISTD% report and not ISTD.
Gasman
Posted: Mon Aug 05, 2013 2:14 pm
by lmh
The calibration graph in ISTD will show amount ratio as one axis, and area ratio as the other axis, and the amounts will be calculated as absolute amounts in whatever unit you gave it. Yes, if you want the amounts to be tabulated in %, definitely follow Gasman's advice.
I didn't answer about replicate standards, and that's because (shame!) I don't know the answer. I've only ever used curves with single vials at each calibration level, often running twice (before and after a bracket of samples), not triplicates, so I've been content with showing the average of the two runs. Of course I always reprocess my standards as though they were samples, so I can see the actual calculated concentrations in both injections of a standard, and check for variation in instrument efficiency, but it's not the same as seeing both as points on the cali curve. The only thing I can suggest is that you add the replicates as separate levels that just happen to contain the same amount of analyte??