Posted: Mon Jul 29, 2013 10:29 am
by stevenbooth
Hi,

I am relatively new to HPLC and I am working with an Agilent 1100 system comprising: Manual injector, quad pump + de-gasser, variable wavelength detector, auto fraction collector.

The issue I am having is; after I have injected the sample chemstation skips the data acquisition phase and jumps immediately to the data analysis phase without collecting any data.

Has anyone experienced this problem before/any ideas for solutions?

Best wishes,
Steven

Re: New to HPLC Chemstation problem

Posted: Tue Jul 30, 2013 2:52 am
by richiekichi
Go to the Method menu, Run Time Checklist, and ensure that Data Acquisition is ticked.