A have a 6890 agilent GC with FID detector, and it is connected to a 7694 HS.
I have to measure benzene from water and detection limit should be under 0.25 ug/l.
We are using N2 as carrier gas and DB WAX column ( 30*530*1)
The parameters are:
HS:
oven temp: 70°
transfer line: 110°
oven equ. : 25 min
shake: low
Vial pressure: 20 psi
GC parameters:
Injector temp: 250°
Oven temp: 35°for 3 min, then 4°/min to 90, 45°/min to 220 and hold for 1 min
FID temp: 300°, H2: 30 ml, Air: 450ml, Makeup (N2) 25 ml
I use 20ml vials, with 10 ml water, about 4g NaCl, with internal std.
As i know, i should use split injection, but i heard spitless can work too. A i have problem with detection limit, i havent been able to reach it yet. I am not sure, if i am using correct flows, from HS, and from GC. Since i am using N2, i need low linear velocity.
Can anybody help me please with it?
Thanks for advance!
Norbi