Drifting Signal 5973
Posted: Fri Jul 19, 2013 8:08 pm
Hello!
I am having trouble with a 6890/5973 which is running P&T being used for 8260. I am hoping to bounce some ideas around with all the expertise here. Problem is as follows; after source maintenance a target BFB tune is preformed and a calibration is run shortly there after. Calibration has excellent response in general. its almost as if the mass spec is too sensitive, low levels have excellent abundances, 4x that of a redundant system running next to it. The high cal points peaks look saturated, they are fat and rounded at the apex, not sharp. this is giving us many quadratic fits in which the high cal point are causing. moving forward with this calibration the MS response fluctuates almost run to run. It most apparent in and overlay of calibration checks (ccv) also in IS responses. The fluctuation is across the entire chormatorgram there is no bias to any specific peak. It can also be seen overlays looking at the TIC baseline.
I have covered many of the typical items troubleshooting this problem;
cleaned source, new filaments, new em (currently tuning 1.5kV), new column, full inlet maintenance, rinsed split lines, new epc, dipped quad per agilent's instruction's at 100 amu, concentrator is a stratum with low miles.
the HED assembly was replaced about 1 year ago, agilent tech support referred to as a noisy hed, I had mass present for every scanned amu. I am thinking there is something still going on... possibly the log amp??
as i am typing I ran the adjust rfpa at amu 800, over ten minutes voltage increased from 554mV to 567mV, is that typical?
any help please! banging my head on the wall right now
. I am happy to elaborate on anything that was left unclear. tried to cover most of the basis.
I am having trouble with a 6890/5973 which is running P&T being used for 8260. I am hoping to bounce some ideas around with all the expertise here. Problem is as follows; after source maintenance a target BFB tune is preformed and a calibration is run shortly there after. Calibration has excellent response in general. its almost as if the mass spec is too sensitive, low levels have excellent abundances, 4x that of a redundant system running next to it. The high cal points peaks look saturated, they are fat and rounded at the apex, not sharp. this is giving us many quadratic fits in which the high cal point are causing. moving forward with this calibration the MS response fluctuates almost run to run. It most apparent in and overlay of calibration checks (ccv) also in IS responses. The fluctuation is across the entire chormatorgram there is no bias to any specific peak. It can also be seen overlays looking at the TIC baseline.
I have covered many of the typical items troubleshooting this problem;
cleaned source, new filaments, new em (currently tuning 1.5kV), new column, full inlet maintenance, rinsed split lines, new epc, dipped quad per agilent's instruction's at 100 amu, concentrator is a stratum with low miles.
the HED assembly was replaced about 1 year ago, agilent tech support referred to as a noisy hed, I had mass present for every scanned amu. I am thinking there is something still going on... possibly the log amp??
as i am typing I ran the adjust rfpa at amu 800, over ten minutes voltage increased from 554mV to 567mV, is that typical?
any help please! banging my head on the wall right now