for a RP-HPLC method can we have auto sampler coller temperature at 20°C and column oven temperature at 35°C. will this resulkt in any peak distortion.
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
Probably not, but the only way to know for sure is to try it.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 9
- Joined: Fri Apr 16, 2010 12:46 pm
I agree with Tom, I have run many related compound HPLC analyses that require the sample be held at 10C, and the column is kept at 40C to speed the analysis. haven't seen any distortion of the peak.
-
- Posts: 12
- Joined: Thu Jun 20, 2013 11:28 am
I agree too.
I set sample Temperature to 20°C and Column to 30°C without problem.
But I wonder if solvent temperature may disturb the run.
Can we see a different between 18°C solvent and 27°C if sample and column temperature are fixed (20 & 30°C)?
I set sample Temperature to 20°C and Column to 30°C without problem.
But I wonder if solvent temperature may disturb the run.
Can we see a different between 18°C solvent and 27°C if sample and column temperature are fixed (20 & 30°C)?
-
- Posts: 3
- Joined: Thu Apr 07, 2011 4:35 pm
Ofcourse you can. Peak distortion would be highly unlikely.
Since the volume injected using the auto-sampler is relatively small (ususally few micro liters), it should not have any influence on the temperature equilibrium the column is already in.
Autosampler is usually kept at a lower temperature to supress any reactions going on in the sample.
Since the volume injected using the auto-sampler is relatively small (ususally few micro liters), it should not have any influence on the temperature equilibrium the column is already in.
Autosampler is usually kept at a lower temperature to supress any reactions going on in the sample.
5 posts
Page 1 of 1
Who is online
In total there are 18 users online :: 0 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 18 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 18 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science