Never-ending problems on the Perkin Elmer FX-15

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
Has anyone experienced continuous problems on the PE FX-15 UHPLC. A brand new machine, one pulse dampener has been replaced... and others....
But right now I am running taurine std's pre-column derivatized by 1-fluoro-2,4 dinitrobenzene to form a UV active analyte. Mobile phase is 80/20 0.01 M phosphate buffer pH 6/ACN. Flow rate has varied from 1 to 2 ml/min from method opt.

The baseline is completely out of whack as you can see Image. What is going on. Thank you so much to everyone.

-Mike
ImageUploaded with ImageShack.com
Hi Mike, I would suggest to increase the mixing volume. As your method is an isocratic method, you can double the mixing volume, maybe 1mL should be OK.
I did this and it worked perfectly.
3 posts Page 1 of 1

Who is online

In total there are 13 users online :: 2 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 11 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry