AutosystemXL Error
Posted: Wed Jul 17, 2013 11:21 pm
by Cyprian
Greetings,
Has anyone ever encountered an error on their Autosystem XL that reads as follows:
"No DTF3 Flow Module at Connection 5.'
Thank you for your assistance.
Cyprian
Re: AutosystemXL Error
Posted: Tue Aug 13, 2013 8:36 pm
by cbwhipp
DTF is the detector flow. It is not recognizing the PPC module (i believe position 5 may be for the air?) on the PPC board. Make sure the module cable connector is seated well on the PPC board. You could have a bad PPC module, PPC board or even a Main board.
Posted: Fri Aug 23, 2013 3:40 pm
by Cyprian
It was the main board along with the PPC module. Thanks for your comment!
Posted: Mon Mar 30, 2015 3:29 pm
by cvggilu
Hi
I have the same issue on my Perkin elmer GC 'Warning CAR 1 pressure less than 50 psi'
Please advice.
Posted: Mon Mar 30, 2015 7:00 pm
by dr_bahram1977
Hi
Please advice.
open your cylinder valve and set it to 100 PSI.
then, check for leaks, and you'll fix it.