Edwards E2M2 foreline pump problem.
Posted: Wed Jul 17, 2013 4:07 pm
by lgchrom
I am using this Edwards pump and it has been fitted with a rebuild kit. Pretty much all consumables are new in it. I used Ultragrade 19 type oil in the pump. It reaches around 230mtorr in operation (the needed spec is around 300). After an hour and a half or so the pressure rises to around 400-450 mtorr. If I let it cool down and re-turn it on it works well again, and then back to the high pressure. I am puzzled. What could be wrong?
thanks in advance
Re: Edwards E2M2 foreline pump problem.
Posted: Thu Jul 18, 2013 12:07 am
by MSCHemist
Should be doing better. I use an E2M1.5 and get down to below 70 mtorr withing a minute or two. Check for leaks. Use a rubber stopper to plug the inlet of the pump, let it go for a minute then see if their is any flow out the exhaust. After the initial evacuation of residual air in the inlet there should be none.
Posted: Thu Jul 18, 2013 1:00 am
by Don_Hilton
After checkign to see if something is leaking - and this can be a foreline connection - either end - if that is not helping you, then:
Was the container of pump oil freshly opened by you? If a pump oil with a higher vapor pressure was mixed into the oil - like put "back" into the wrong container, you can have a problem obtaining the vacuum you need when the pump warms up...
Posted: Fri Jul 19, 2013 5:46 pm
by jvettleson
We typically use Inland 45 oil with these types of pumps. Can't speak to their performance with the oil you have in it.
Posted: Fri Jul 19, 2013 8:47 pm
by Don_Hilton
Was there a type of oil specified for your system? Properties of oils do differ note:
http://www.inlandvacuum.com/index_files/mechanical.htm and
http://www.tedpella.com/vacuum_html/Vac ... erties.htm The vapor pressure at operating temperature is the critical item, by the way - and it would take further digging to come up with the vapor pressure of inland 45 at an elevated temperature.
Posted: Mon Jul 29, 2013 2:09 am
by mckrause
I agree with MSChemist. You should be seeing WAY better than 230 mTorr with an E2M2 on a GC-MS. If you are using it on an LC-MS then you desperately need to change pumps, as the E2M2 is badly undersized for a LC-MS.
If you are on a GC-MS then you have what amounts to a massive leak somewhere in your vacuum system.
Posted: Mon Jul 29, 2013 9:18 pm
by James_Ball
We always use Inland45, and from that chart the vapor pressure @25c has a difference of 1x10^-5 torr for Inland19 versus 1X10^-7torr for Inland45. I would think his Ultragrade 19 oil would be similar to the Inland19.
Try some Inland45 in the pump and see what happens. I get 70mtorr or less with 1ml/min flow through the MS on my E2M2 once it is warmed up.
Posted: Mon Aug 19, 2013 6:26 pm
by Rick
I had this problem in 2003.
The shaft seal is probably not properly install......not the shaft seal near motor but the other.