Detector gain calibration?
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
2 posts
Page 1 of 1
I have a Thermo DSQ GC/MS that I am trying to tune and it keeps failing the detector gain calibration. What is this even for? And how can I make this pass? I put in a new PMT and it seemed to help he tune I did right after that, now 2 weeks later it is failing again. Website, book, explaination... anything, please help me understand!
-
- Posts: 9
- Joined: Mon Feb 14, 2011 1:05 pm
Detector gain calibration adjusts the voltage to the detector to reach a spacific gain. It can be unchecked in the Autotune menu. It is not necessary to be run as often. If it fails probably the ionisation is not good and intensity of ion 100 is low. Did you clean the source? Did you change the filament? Is the column installed correctly? Changing detector is last resort. I believe the DSQ uses electron multiplier not a photomultiplier, but may be there are some like that.
2 posts
Page 1 of 1
Who is online
In total there are 8 users online :: 0 registered, 0 hidden and 8 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 8 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 8 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science