New to HPLC Chemstation problem

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Hi,

I am relatively new to HPLC and I am working with an Agilent 1100 system comprising: Manual injector, quad pump + de-gasser, variable wavelength detector, auto fraction collector.

The issue I am having is; after I have injected the sample chemstation skips the data acquisition phase and jumps immediately to the data analysis phase without collecting any data.

Has anyone experienced this problem before/any ideas for solutions?

Best wishes,
Steven
Go to the Method menu, Run Time Checklist, and ensure that Data Acquisition is ticked.