We just recently had a similar issue on an Agilent 1200 system essentially the same, but this had DAD detector and wavelength of 280nm.
First we isolated that it was NOT the pump (we set flow to zero or shut off pump). The repeating pattern remained, sure seemed like it was detector-related, not related to flow at all, did not change. Then we turned pump on but slid the detector flow cell out, still repeated.
Unfortunately, we do not yet have Lab Advisor software, so we couldn't run a lamp diagnostic test, or even see how many hours the lamp had been on in its lifetime (the system came from a different department).
So I installed a new old-stock UV lamp (from May 2008, never used), ran the system overnight, and I've since run another sequence and no issues.
So - for us - so far it seems like it was an old lamp. Notice that I said "so far".