lmh » Mon Jul 08, 2013 9:34 am
I'm struggling with an elderly Surveyor MS pump (Thermo). It's started to have what looks like check-valve problems (pistons move, no pressure, if I open the purge line, liquid goes in and out; same on all lines, so it can't be a GPV issue, happens in mid sequence when pressure has previously been stable, so I don't think it's a sudden bubble; forcing liquid forwards through pump usually frees it, and it starts pumping OK again, but not for ever...). This has happened at a time someone is pumping fairly high rates of 100% acetonitrile to degunk their column after making injections of samples that contain something hydrophobic that otherwise accumulates on column. Failures tend to happen correlating with their 100% acetonitrile clean-up runs. But the pump is also 10 years old... (although obviously maintained, with regularly-changed check-valves; this problem has happened on 2 sets of check-valves, one new, one of dubious history).
(1) Does anyone else have experience of this sort of problem on that pump?
(2) Is it likely to be an acetonitrile-with-check-valves problem?
(3) If so, is it possible to use a mix of acetonitrile and methanol as an alternative strong organic clean-up, without check-valve problems?
Thanks in advance for any comments and help.