Help,

I am trying to figure out how to use chromeleon to make a program with a scanning wavelength. I am running on an Agilent 1100 with a DAD. Where do I start? I have tried with the help index and have had no luck...is it impossible?

Thanks for any and all help,

Tim
Hi, after the instrumentconfiguration is done, you do as for any other diod-array detector. No difference in the program itself. That is the nice thing with Chromeleon.
Hi Morris,

The first thing to check is that you have the required 3D license to allow full spectral scans to be performed. If you don't have this, then you will not be able to do spectral analysis.

However, if you do have this license, you will just need to set the appropriate wavelength range within the Program File (PGM). How you do this will depend upon whether you are using Agilents ICF (Instrument Control Framework) driver, or the native Chromeleon Driver within your Server Configuration.
Anthony

ThermoFisher Scientific
3 posts Page 1 of 1

Who is online

In total there are 19 users online :: 0 registered, 0 hidden and 19 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 19 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry