Atlantis T3 fast efficiency loss
Posted: Mon Jul 08, 2013 11:30 am
Hi, I have a problem with my Atlantis T3. I have used Atlantis T3 3um 2.1x100mm with UPLC system coupled with mass spectrometer in order to identify some compounds. After that analysis was done using HPLC system. In order to maintain similar gradient elution conditions I bought Atlantis T3 3um 4.6x150 mm column (bought longer column to achieve better separation in comparison with 4.6x100mm). First chromatogram was very similar to UPLC, peaks were not identical but separation was very good. But after 30-40 injections the efficiency was very low and peaks were very wide - efficiency become terrible, but retention times were the same.
I used two solvents: A: 10mM TEAA pH7 and B: ACN. Gradient conditions: -10min 95 A and 5 B, 0min 95 A and 5 B, 11min 0 A and 100 B, 16min 0 A and 100 B (linear gradient).
After some injections I washed column 95 H2O and 5 ACN for 15 min (in order to eliminate TEAA from column) and then 100 ACN for about 15 min. After washing loss of efficiency was observed. But efficiency declined from inject to inject, too.
I guess the problem could be too low ACN conc. in washing eluent or the part of gradient where 100 ACN is used for 5 min (11-16) caused efficiency loss (maybe triethylammonium acetate precipitates). System backpressure is stable and analytes in the mixture is stable too.
I would like to read other opinions and suggestions. Maybe someone used Atlantis columns and could help me with proper washing and equilibrating conditions of this column.
Sorry for my English, it is not my mother tongue.
I used two solvents: A: 10mM TEAA pH7 and B: ACN. Gradient conditions: -10min 95 A and 5 B, 0min 95 A and 5 B, 11min 0 A and 100 B, 16min 0 A and 100 B (linear gradient).
After some injections I washed column 95 H2O and 5 ACN for 15 min (in order to eliminate TEAA from column) and then 100 ACN for about 15 min. After washing loss of efficiency was observed. But efficiency declined from inject to inject, too.
I guess the problem could be too low ACN conc. in washing eluent or the part of gradient where 100 ACN is used for 5 min (11-16) caused efficiency loss (maybe triethylammonium acetate precipitates). System backpressure is stable and analytes in the mixture is stable too.
I would like to read other opinions and suggestions. Maybe someone used Atlantis columns and could help me with proper washing and equilibrating conditions of this column.
Sorry for my English, it is not my mother tongue.