xxx13 » Tue Jul 09, 2013 9:14 pm
													
 
						
					 
					
						So, the only new component that lead to the cleaned up background was a new needle?  You stated previously that a new capillary (what instrument are you using, by the way?) decreased but did not eliminate the noise.
I would try to place the old capillary and needle back in the instrument, just to see if this is something directly associated with those parts, or if this was just some in-source contamination that needed to bake out to reduce to undetectable levels.
Another thought - I saw an article about siloxane contamination in nanoESI, and they were able to reduce the level of interference by increasing nitrogen flow around the needle and repositioning the needle in the source.  Maybe in replacing the needle you inadvertently repositioned it and reduced the interfering signals to undetectable levels?  Maybe you can try different needle positions to see if the contamination comes back?
http://www.ncbi.nlm.nih.gov/pubmed/12794873
Thermo Finnigan LTQ
step 1: disconnect LC from MS, the high background is still there, so it is not associated with LC.
step 2: sonicate the needle and cone again, problem remaining;
step 3: the thermo engeneer adviced to replace a new capillary, I did it, the noise decreased a little and still high background;
step 4: the thermo engeneer adviced to replace a new needle, I did it, the action does not work;
step 5: infusing calibration mixture using tune function, all MS parameters are normal;
step 6: increase nitrogen flow rate, it does not eliminate the problem;
I am not sure if I inadvertently repositioned the ESI probe while changing the needle or installing the probe back. Because the noise is dramatically increasing, the intensity and resolution decreased for real samples. The needle is fixed on the ESI probe, the needle tip would vary the position in a small 3D space.