4-nitrophenyl 2-chloroethylcarbamate ,No signal by MS detect

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

8 posts Page 1 of 1
Hello,everyone:

I detect 4-nitrophenyl 2-chloroethylcarbamate with LC-MS.but I cound not find the peak with m/z=245.what wrong with me ?

The test condition is :

Instrument: thermo LTQ-ORBITRAP, ms detection is LTQ;
Scan mode :"+"
ESI heater temp:300degree
I spray voltage: 3KV
capillary temp:300degree
scan range(m/z): 50-500

mobile phase contain : 0.1% formic acid. or 10mmol/L ammonium acetate

Sorry, I cann't paste the chemical structure in this page :?:

thanks of all
When using the Ammonium Acetate in the Mobile Phase you may be generating the M+Ammonium adduct. Try looking for M+H+17 as the mass(m/z=262).
The past is there to guide us into the future, not to dwell in.
When using the Ammonium Acetate in the Mobile Phase you may be generating the M+Ammonium adduct. Try looking for M+H+17 as the mass(m/z=262).
Thanks, James_ball,I look for the peak with m/z=262,there is no.Now I am trying to find if there is some of its fragments with +Ammonium adduct.
Thank you.
Are you using a column (you mention mobile phase)? If so, are you sure it is eluting? Might try running it straight into the source at relatively high concentrations to get a true spectra, at that way you know wht to look for under your conditions.
Mark Krause
Laboratory Director
Krause Analytical
Austin, TX USA
Are you using a column (you mention mobile phase)? If so, are you sure it is eluting? Might try running it straight into the source at relatively high concentrations to get a true spectra, at that way you know wht to look for under your conditions.
Dear sir/madam:
I detected the compound by ms and lc-ms.there are all no 245[M+H ]or 262[M+NH4].I want to know if the compound has signal in MS?
Suggest that you search the mass spectrum for any signal that shows the characteristic 3:1 signal for an ion that contains a single Cl atom.

This will show you whether the analyte has degraded in the source.

What was your scan range ??
MS/MS on the [M+H]+ ion would be expected to give, inter alia,

MH+ - HCl; m/z 209 (singlet)
MH+ - NO2; m/z 199/201 doublet

NO2C6H4O-CO+; m/z 166 (singlet)
CH2CH2Cl+; m/z 63/65 doublet.

If the molecule is unstable you might see some of these under non-CID conditions.
Back to basics---do you have an HPLC/UV chromatogram to show that the analyte was actually eluted from the column ??
8 posts Page 1 of 1

Who is online

In total there are 22 users online :: 0 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 22 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry