Hi Dory,
So it's a Shodex RI detector whose celll is maintained at 35 degrees Celsius, a Thermo Dionex UHPLC (or is this an analytical scale instrument?) whose column heater is kept at 18 degrees Celsius and a flow rate of 0.4 mL/min of 10 mM sulfuric acid, and apparently these days one can employ acetonitrile, IPA and ethanol at certain levels and times, although not methanol, ever, with the Bio-Rad Aminex HPX-87H. Got it.
A few questions:
Why do you run your sucrose separations at a temperature so much lower (18 degrees Celsius) than the recommended values (50 - 65 degrees Celsius) for the Aminex HPX-87H?
At this low temp and flow rate, what is your backpressure reading? There's kind of a strict limit on the backpressure these columns can tolerate, 1500 psi as I recall.
What type of injector does the Ultimate 3000 employ, flow-through needle or fixed loop?
What type of injections are you making, full-loop or partial-loop?
If you exchanged columns, agreed, a possibility is a dirty injector. Are you using some kind of seal wash that contains no methanol and/or a needle wash--I've not the foggiest how the Ultimate 3000 works with respect to these types of solutions. Looks like you can remove the column and purge the injector, and perhaps the sample loop as well, with the Ultimate 3000, most likely warm water (40 degrees Celsius) would be fine for this, to remove sucrose.
Probably an entire system flush with warm water containing 10-15 percent by volume acetonitrile is also in order, again without the column in place.
Please, see what you think.