Agilent 1200, Chromeleon 7.1 and Corona Ultra detector

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

3 posts Page 1 of 1
Got the Agilent 1200 set up through ICF, Chromeleon 7.1, have control of the instrument. At first put the CAD under the same instrument and could not see the CAD, set up another instrument with just the CAD under the same controller. The CAD is connected via USB. The Agilent is using Jet Direct card - to dedicated NIC card.

Once CAD set up as a separate instrument- all menu items are visible, however there is not trace, nothing appears when monitoring baseline. I have the CAD under remote control and it is responding to commands from the menu. After a sequence is setup the following error message shows up:

"No data has been received from the detector for 2 minutes."

when CAD is setup as stand alone it appears to want to be connected to Ultimate 3000 in chromeleon, however this instrument does not exist.

Very close to getting this working but really without a signal- I am not sure I can use the software with current configuration.
Hi Chrom,

Firstly, have you tried contacting your local Tech Support? They may be better placed to support you with this.

However...

Within the Configuration Manager, just a few of things to check.

1) You have 'Simulation Mode' un-checked and are therefore connected to your real instrument?
2) You have browsed to the USB module address to allow correct connection?
3) Within the Detectors Tab, do you have the right version of your CAD selected? i.e. Corona, Corona Plus, Corona Ultra or Corona Ultra RS.

In reality, there shouldn't be any issues with what you are trying to do.

Thanks
Anthony

ThermoFisher Scientific
Hi Chrom,

Firstly, have you tried contacting your local Tech Support? They may be better placed to support you with this.

However...

Within the Configuration Manager, just a few of things to check.

1) You have 'Simulation Mode' un-checked and are therefore connected to your real instrument?
2) You have browsed to the USB module address to allow correct connection?
3) Within the Detectors Tab, do you have the right version of your CAD selected? i.e. Corona, Corona Plus, Corona Ultra or Corona Ultra RS.

In reality, there shouldn't be any issues with what you are trying to do.

Thanks
Hi, everything has been set up properly, the CAD is not in simulation mode, but after 2 minutes it will not collect data and therefore stops communicating. We are not really local to tech support, so I called to get a service contract on the system since it has sat idle for 2 years and after talking with a few contacts from TMO I believe the board may be the issue.

Chrom.
3 posts Page 1 of 1

Who is online

In total there are 17 users online :: 1 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google Adsense [Bot] and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry