Problem in negative ionistion in Quottro Micro LCMSMS
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
8 posts
Page 1 of 1
We have a problem with our waters Quottro Micro LCMSMS.we have operating in ESI mode.We have not getting peeks of negative ionisation componds in multimode ionisation modes.We have tried in single ionisation mode with only negative ionisation also but we have not getting peeks of negative ionisation compounds.What is the cause of this?
-
- Posts: 762
- Joined: Thu Dec 13, 2007 5:54 am
Has this analysis worked before? What specific compounds are you looking for in both positive and negative ionization? What is your mobile phase? By your limited description, I take it you also have compounds that respond with positive ionization, and they are responding properly?
Time flies like an arrow. Fruit flies like a banana.
-
- Posts: 61
- Joined: Fri Sep 02, 2011 1:26 am
yes this analysis is worked nicely earlier.we have this problem starting one week earlier.We analysing compounds like chloramphenicol,TFNA,fipronil etc in -ive ionisation.The positve ionisation componds are ok.
-
- Posts: 762
- Joined: Thu Dec 13, 2007 5:54 am
OK, what mobile phase are you using?
Time flies like an arrow. Fruit flies like a banana.
-
- Posts: 61
- Joined: Fri Sep 02, 2011 1:26 am
we are using acetonitrile and water in single iomisation mode for chloramphenicol and 5mm Ammonium formate in water and Methanol for Mixmode ionisation.
-
- Posts: 762
- Joined: Thu Dec 13, 2007 5:54 am
Have you done the usual - re-make your mobile phases, purge the LC and degasser of the old mobile phase, and re-make your chloramphenicol standard? Did anything occur a week ago, like running a different analysis, or maintenance?
Time flies like an arrow. Fruit flies like a banana.
-
- Posts: 67
- Joined: Tue Nov 11, 2008 4:18 pm
Hi Davidantony,
As I wrote in a previous thread regarding loss of sensitivity in a QMicro it could be due to a dirty first quadrupole.
As I wrote in a previous thread regarding loss of sensitivity in a QMicro it could be due to a dirty first quadrupole.
Mike
-
- Posts: 61
- Joined: Fri Sep 02, 2011 1:26 am
we have cleaned the source and 1 st quadrapole.Then make some adjustments. now the response of the ngative and positive compounds is good.
8 posts
Page 1 of 1
Who is online
In total there are 24 users online :: 1 registered, 0 hidden and 23 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot] and 23 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot] and 23 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science