Hi all,

I just have this question relating the abundance of a GC/MS. Let's say I have a new GC/MS instrument, with certain compound with the same concentration, I always get a signal of one million counts or similar. After half year or so, the abundance gets lower, so I clean the source and do the autotune (of course the EMV is higher : from 1300 to 1700). I expect the compound signal still gets around one million. But it is only 700k or so. Is there anything wrong ?
Thanks for any inputs.
I do not believe an autotune can achieve that level of precision. I would simply push the EMV up a bit from the autotune (or whatever type of tune process you like) to get a similar area. Sometimes this is a meaningless act, as the S/N may stay the same.

Keep in mind the tuning process is using a compound (often PFTBA) that is in rough vapor pressure equilibrium with the vacuum. Not really a precise setup.

I usually try to achieve a response that is within 50 percent of whatever I was getting when the MDL, LOD, etc was determined, biased towards the high side. If I get at least 2 years out of a horn, I do not mind replacing it.
I do not believe an autotune can achieve that level of precision. I would simply push the EMV up a bit from the autotune (or whatever type of tune process you like) to get a similar area. Sometimes this is a meaningless act, as the S/N may stay the same.

Keep in mind the tuning process is using a compound (often PFTBA) that is in rough vapor pressure equilibrium with the vacuum. Not really a precise setup.

I usually try to achieve a response that is within 50 percent of whatever I was getting when the MDL, LOD, etc was determined, biased towards the high side. If I get at least 2 years out of a horn, I do not mind replacing it.
Thanks
I agree with Yama, best way is to do the autotune then inject a standard and manually adjust the EMV up or down to bring the area counts back to near what you originally had. I often step it up a few volts when needed to keep the areas up between cleanings.
The past is there to guide us into the future, not to dwell in.
I find you can't really tell to much from the abundances except that if they are low you have a big problem. I had a cracked repeller insulator causing a 90% drop in signal and the abundances looked fine on the tune.
5 posts Page 1 of 1

Who is online

In total there are 21 users online :: 1 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Ahrefs [Bot] and 20 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry